Wednesday, November 23, 2016

experimental physics - Photomultiplier/ voltage divider troubleshooting


I have a Photonis XP5301 PMT to use for fast neutron detection (detector optically glued to the PMT window). I can't quite figure out why its not producing a spectrum. First of all, I have multiple different values of operating voltage that I found on the internet, but I think it's either -800 V or - 1000V. I used the same setup with a NaITl scintillation detector + attached PMT, and it gave me proper results, but not mine. What could I do to locate the issue? The Voltage divider I'm using is a Photonis VD202K/40. ALso, while the NaITl detector's PMT has a power cable that connects to the amplifier, my PMT doesnt have a power cable. What could be the issue here? What am I not addressing? Thank you!



Answer



You say you are using the setup for a neutron detector.


What is the material you are using? What is the expected light output (in photons / event) and how fast would they appear? When you use the NaI based detector, what is your source? What is the activity?


Here are the troubleshooting steps I would take:



  1. To make sure the PMT is working, I would hook the output (straight from the PMT anode) up to a scope, into a low-ish impedance (1 k to ground). Turn the gain of the scope to maximum (or 10 mV / cm). Slowly increase the operating voltage (while PMT is in darkness); eventually, you should see "grass" on the baseline of the scope: this is the signal from individual (thermally induced) photoelectrons.


  2. Turn the voltage down about 100 V from the point where noise started appearing, and introduce a very weak flashing LED into the enclosure of the PMT (if you can see the flash with the naked eye it's too bright for the PMT). You should now see a massive spike of signal for every flash of the LED. Turn the HV off and demonstrate to yourself that the signal is gone (so there is no cross talk)

  3. Now connect the PMT output to a shaping amplifier (it's not clear whether you have one - if you don't, then that might be your problem). The shaping amplifier should give a "slow" output at every pulse coming in - this is usually needed as the input to a spectrum analyzer. Check the output of the amplifier on your scope (you will need a much less sensitive setting). Again, using the LED could be very helpful here. If you don't get a signal from the amplifier, check its power etc.

  4. Now that you have a PMT with roughly appropriate voltage and a working amplifier, connect to the spectrum analyzer. Introduce a robust source (so you get good signal) and slowly sweep the HV in the range you found above, until you see triggers on the spectrum


When you have done all the above, and you found a place where things don't work as expected, you've likely narrowed your problem significantly. If you still have trouble, leave a comment.


No comments:

Post a Comment

classical mechanics - Moment of a force about a given axis (Torque) - Scalar or vectorial?

I am studying Statics and saw that: The moment of a force about a given axis (or Torque) is defined by the equation: $M_X = (\vec r \times \...